PL33 motorised probe head cable
| Cable name | Part number | Length |
| PL33 | A-1023-7056 | 3 m (118.11 in) |
This straight cable has in-line LEMO socket or plug alternatives supplied loose for fitment on end 'A'. A 90° Tuchel socket is already fitted to end 'B'. This cable electrically connects Renishaw PH10T, PH10M or PH10MQ motorised probe heads to the CMM quill.
End 'A' plug | End 'A' socket | End 'B' | Wire | Signal | Max. line current |
1 | 3 | E | Yellow | 'B' axis feedback | n/a |
2 | 2 | D | Red | Ground sense | n/a |
3 | 1 | C | Brown | dc reference 12 V | n/a |
4* | 7Ŧ | M | Grey | 0 V | 1000 mA |
5 | 6 | H | White | Locking motor 8 V dc nominal | 350 mA |
6 | 5 | L | White / black | 'A' axis motor 12 V dc nominal | 350 mA |
7* | 4Ŧ | n/a | Not connected | Head present | n/a |
8 | 11 | F | Blue | 'A' axis motor 12 V dc nominal | 350 mA |
9 | 10 | A | Violet | 'B' axis motor / probe contact | 350 mA |
10 | 9 | B | Black | 'B' axis motor / probe contact | 350 mA |
11 | 8 | N and O | Screen | Screen | n/a |
12 | 14 | G | Orange | 'A' axis feedback | n/a |
13 | 13 | J | White / brown | LED and datum | 15 mA |
14 | 12 | K | Green | Motor probe switch | 40 mA |
* NOTE: The male pins 4 and 7 of the 14 way LEMO plug (end 'A) are linked together.