Safety recommendations for the UCCFusion controller
Environmental conditions
Indoor useIP30* (BS EN60529:1992)
Altitudeup to 2000 m
Operating temperature+5 °C to +50 °C
Storage temperature-25 °C to +70 °C
Relative humidity
80% maximum (non-condensing) for temperatures up to +31 °C
Linear decrease to 50% at +40 °C
Transient voltagesInstallation category II
Pollution degree2
NOTE: It may be necessary to house UCCFusion in a suitable electrical enclosure according to the installation's environmental conditions to obtain a higher IP rating.
NOTE: For ambient temperatures under 40 degrees Celsius, maximum input power for the UCCFusion is 720W. Temperatures between 40 and 50 degrees Celsius the maximum power is 400W.
WARNING: UCCFusion is not recommended for use in ambient temperatures above 50 degrees Celsius.
WARNING:  Earthing provisions for the complete machine is the responsibility of the OEM or installer.
WARNING: Isolation: UCCFusion is isolated from the DC power supply by disconnection of the power connector. If any additional means of isolation is required, it must be specified and fitted by the machine manufacturer or the installer of the product. The isolator must be sited within easy reach of the CMM operator and comply with IEC61010 and any applicable national wiring regulations for the country of installation.
WARNING: The PSU must be single fault tolerant, approved to EN60950-1 and must be connected to a supply incorporating a protective earth conductor. Earthing provisions and residual fault protection for the complete machine is the responsibility of the OEM or installer.
Machine safety
WARNING: Switching off or isolating the UCCFusion may NOT prevent unexpected machine movement. The user is advised to isolate the machine from the electricity supply, compressed air or other energy sources in accordance with the machine manufacturer's instructions before entering the danger zone or performing any maintenance operations.
If the equipment is used in a manner not specified by the manufacturer, the protection provided by the equipment may be impaired.
The emergency stop safety system that has been integrated into the UCCFusion controller is designed to achieve category 2 to standard BS EN ISO13849-1:2008, this can be hardware configured to obtain category B. The risk assessment made by the system installer on the installation should therefore have determined which category is required.
The UCCFusion must only be commissioned through Renishaw software, failure to do so may result in unexpected machine movement.