AC1 software interface
Board identification (base address + 15)
This read-only register returns the value 0DH, which indicates the presence of the AC1 card.
Status register (base address + 14)
AC1 I/O map
Register name
Base address offset
Bit reference
Data
Read/write
X axis LO byte
0
7 to 0
2’s complement 12 bit count 8000H to 7FFFH
 Read only
X axis HI byte
1
15 to 8
 Read only
Y axis LO byte
2
7 to 0
 Read only
Y axis HI byte
3
15 to 8
 Read only
Z axis LO byte
4
7 to 0
 Read only
Z axis HI byte
5
15 to 8
 Read only
Timer LO count
6
7 to 0
16 bit binary count 0000H to FFFFH
 Read only
Timer HI count
7
15 to 8
 Read only
Not used
8-12
N/A
 N/A
 N/A
Command register
13
15 to 8
 Relevant bit set to logic 1 to perform command
 Read only
Status register
14
7 to 0
 Bits set to logic 1 if condition TRUE
 Read only
15
15 to 8
 8-bit board identification
 Read only
Command register definitions
15
Not used
14
Not used
13
Not used
12
Not used
11
ACQUIRE data and latch timer count (sets BUSY true until complete)
10
REQUEST SET PROBE PRESENT
9
REQUEST RESET OVERTRAVEL
8
RESET TIMER
Status register definitions
7
Not used (set to logic 1)
6
BUSY
5
TIMER OVERFLOW
4
PROBE PRESENT
3
OVERTRAVEL
2
5 V FUSE BLOWN
1
-12 V FUSE BLOWN
0
+12 V FUSE BLOWN
This read-only register returns the status of the AC1 card.
AC1 status bits
BIT 7
Not used
BIT 6 - BUSY
This bit is set to “1” when the AC1 has been commanded to acquire data. It is reset to “0” when the data conversion is complete. This takes approximately 85 μs.
NOTE: The host PC should not attempt to read the probe axis deflections or the timer value while the BUSY bit is set to “1”. Any data read during this period will be invalid.
Bit 5 - TIMER OVERFLOW
This bit is set to “1” when the timer has overflowed. It is reset to “0” when the timer has been reset.
Bit 4 - PROBE PRESENT
This bit is set to “1” if an SP600/M/Q probe has been connected to the AC1 and the REQUEST SET PROBE PRESENT bit has been written to with a “1” (table in section Command register definitions earlier on this page). If the probe is disconnected from the AC1, then the bit automatically resets to “0”.
Bit 10 - REQUEST SET PROBE PRESENT
Writing a “1” to this bit causes the AC1 to sample the state of its probe identification circuitry. Upon writing to this bit, the PROBE PRESENT bit in the status register is set to “1” if an SP600/M/Q probe is present.
Bit 9 - REQUEST RESET OVERTRAVEL
Writing a “1” to this bit causes the AC1 to sample the state of its overtravel circuitry. Upon writing to this bit, the OVERTRAVEL bit in the status register is set to “0” if an overtravel unit is connected and
not overtravelled.
Bit 8 - RESET TIMER
Writing a “1” to this bit causes the AC1 to reset the AC1 timer to 0000H and resets the TIMER OVERFLOW flag to “0”.
Timer count LO and HI bytes (base address +7 and +6)
These read-only registers return the latched LO an HI byte of the AC1internal timer. The latched timer value is updated each time the axis defl ections are acquired.
NOTE: It is possible for the counter to have an offset of up to 1 μs from the time of writing to the RESET TIMER bit.
Axis deflection LO and HI bytes (base address +5 through to 0)
These read-only registers return the LO and HI byte of the X, Y and Z axis defl ections of the probe. The format of these values is 2’s complement 12 bit counts which can take any value between 8000
Hex and 7FF0 Hex. The defl ections are only updated when the ACQUIRE bit is written to.
NOTE: As the axis deflections are 12 bits long and the axis deflections are read in two 8 bit bytes, the data has been shifted to the left of the data word. The least significant 4 bits of the lower byte are always zero.
