This section on troubleshooting is a guide to problems associated with the installation and integration of the system only. Refer to the 'PH10 PLUS installation and user's guide' (Renishaw part number H-1000-7592) regarding problems associated with normal operation of the system.
The optional HCU2 can also be used to identify system faults. For full details of the use and fault finding capability of the unit see the 'HCU2 hand control unit user's guide' (Renishaw part number H-1000-5361).
Use the table below to identify problems you are experiencing with your system. If you experience problems which you are not able to identify or solve satisfactorily, please contact Renishaw for further advice.
Power on LED not lit
Possible cause
Power loss
Check mains cable connections and integrity.
Check power block is supplying 24 V.
Check CMM emergency stop condition - power may have been removed by CMM.
No head movement in automatic mode
Possible cause
Head controller in manual mode.
Change to automatic mode.
Power loss
Cable / connection fault
Check connection and integrity of cabling between head and controller.
Communications failure
Check RS232 baud rate.
Incorrect command set selected
Check correct command set is selected.
PICS STOP is asserted by other system component
Clear the fault asserting PICS STOP.
No head movement in manual mode
Possible cause
Head controller in automatic mode.
Change to manual mode.
Power loss
No probe output signal received by the CMM computer
Possible cause
Check connection and integrity of cabling between controller and interface.
Check connection and integrity of cabling between interface and computer.
Multiwire bypass connector not fitted
Fit a multiwire bypass connector with will permit a standard touch-trigger probe signal to reach controller
Multiwired probe in use
Check that the multiwire cable is correctly fitted to the head.
Check that the trigger output to the CMM computer is connected to the multiwired probe interface.
Poor measurement performance
Possible cause
Loose head mounting
Ensure all mounting screws are tight and that mounting to the CMM is secure.
Probe incorrectly attached
Remove probe and reattach.
Force applied to head during lock up
Lock and unlock the head.
Incorrect lock up position
Reposition head correctly.
Position not qualified
Check qualification information.
Probe damping enabled during measurement
Ensure probe damping is not enabled during measurement moves.