![]() | PH10 Motorised Probe Head SpecificationAll of the information below is valid for PH10T/M/MQ-iQ Environmental conditionsThe following environmental conditions comply with those defined in BS EN61010 - 1:2001:
Repeatability of position | |
PH10T PLUS | <0.4 μm (0.00002 in) equivalent to a TP20 and stylus (part number A-5000-3604) |
PH10M PLUS / PH10MQ PLUS | <0.4 μm (0.00002 in) equivalent to a TP6A and stylus (part number A-5000-3553) |
Accuracy of step spacing from theoretical position | |
PH10T PLUS | ±0.24 mm (±0.012 in) using TP6 probe and stylus (part number A-5000-3553) |
PH10M PLUS / PH10MQ PLUS | ±0.3 mm (±0.012 in) using TP6A probe and stylus (part number A-5000-3553) |
Positioning | |
Step | 7.5° |
A-axis 0° to 105° | 15 positions |
B-axis -180° to +180° | 48 positions |
Total | 720 positions |
Probe mounting | |
PH10T PLUS | M8 thread, 18 mm diameter bush face |
PH10M PLUS / PH10MQ PLUS | Multiwired autojoint |
Dimensions from quill face | |
XY (all) | 62 mm |
Z (PH10T PLUS) | 102 mm |
Z (PH10M PLUS) | 117 mm |
Z (PH10MQ PLUS) | 73 mm |
Weight | |
PH10T PLUS | 595 g |
PH10M PLUS | 620 g |
PH10MQ PLUS | 730 g |
System operating temperature | +10 °C to +40 °C (+50 °F to +104 °F) |
System storage temperature | -10 °C to +70 °C (+14 °F to +158 °F) |
The PHC10-3 PLUS is powered from the ac mains supply via an IEC 320 connector.
The operating voltages of the unit are as follows:
100 Vac - 240 Vac +10%, -15% 47 Hz - 66 Hz, 30 W maximum